MEMS Sensor Systems and IoT Part I(cavity)Cavitas biosensors using biocompatible
polymer and MEMS techniques
Kohji MitsubayashiInstitute of Biomaterials and Bioengineering TOKYO MEDICAL and DENTAL UNIVERSITY
MEM Engineer Forum 2017April 26 - 27, 2017, KFC Hall Ryogoku, Tokyo, Japan
Outline Concept of “Cavitas sensors” for non-invasive
bio/chemical monitoring as daily medicine.
Soft contact-lens type biosensorusing Soft-MEMS techniques.# Fabrication w/ biocompatible polymers# Evaluation of sensor behavior# Tear sugar monitoring
Mouth guard type biosensor with BLE 4.0 telemetric system
Wearable sensor tech.
Transcutaneous O2 (tcPO2) sensor
Oraganic + MEMS
Sensors and Actuators B 95: 373–377, 2003Sensors and Actuators B 108: 733–737, 2005Biomed Microdevices 9:1–6, 2007
Opt. transparent
[1] Soft-CL Biosensor
Tear sugar analysis
[2] Body Area Network
Biomed Microdevices, 11:837–842, 2009Talanta 83: 960–965, 2011Biomed Microdevices (2011) 13:603–611, 2011Materials Transduction, 53: 1657-1664, 2012Analytical Letters, 47: 1361–1374, 2014
Biosensors and Bioelectronics 22: 558–562, 2006Biomed Microdevices 9:603–609, 2007Anal Bioanal Chem 391:1269–1274, 2008Photopolymer Sci. 23: 167-170, 2010Materials Transactions, 52: 1943-1948, 2011
Anal. Sci. 17: 773-775, 2001Biosensors and Bioelectronics 19: 67/71, 2003
Wearable bio/chemical sensors and Sensor communication system
Self analysis
Portable inst.physical & mental sufferinginfection
component Contained amount (mg/dl)
ascorbic acid 0.14citric acid 0.60urea 0.04ammonium 5.00glucose 2.60 (0.14 mM)amino acid 7.58
Components of tear fluids
Investigation of SCL type biosensor with flexible sensor techniques
Biological Information on Eye Site
diabetic patient6.9 million in Japan
15.7 million in USA
diabetic retinopathydiabetic nephrosclerosis
(intensity)
5 min time-delay
disease complication
Soft Contact lens (CL) glucose sensor
Requirements• Optical transparency • Flexibility & mechanical strength• Utility as a contact lens
1. Sensor fabrication
• Design & fabrication• Electrode deposition• Characterization
2. Material investigation
• Biocompatibility • Polymer synthesis• Enzyme immobilization
H2O2 (enzymatic product)
glucose
GODglucose + O2 gluconic acid + H2O2
Pt
Ag/AgCl
+400 mv vs. Ag/AgCl
(a) PDMS to lower-template
(b) CL with upper-template
(c) Drying
(d) Release from template
Soft-CL electrode construction with polymer and MEMS
sputtering
Mask sputter deposition
Ti stencil
Ti stencil fabricationLaser marking system
poly-dimethyl siloxane
Sputtering method Formed film on PDMS
Ion Beam Sputtering
Parallel Plate Sputtering Cracks and pealing off
Sufficient adhesion!
Sputtering method on PDMS sheet
Soft-CL type electrode
glucose + O2 gluconic acid + H2O2GOD
enzyme reaction
PMEH by radical polymerizationIII. Dialysis IV. Re-precipitation
PMEH
II. Radical polymerization
V. Dryingdialysis tube
reflux condenser
three-neckedflask
oil bath
solvent (80wt%)
polymerization initiator
Ar bubbling(deoxidization)
I. Preparation of monomer
monomer (20wt%) drying tube
Removal of residual monomerby agitation with stirrer collect rate = 27.2 [%]
Enzyme immobilization on soft-CL Biocompatible MPC polymerOrganic synthesis of biocompatible polymer for Soft-CL biosensor
without any anesthetic
Application of SCL sensor on Rabbit eye siteExperimental set-up for continuous monitoring of tear glucose
sex:male
age:7 months
weight:3.6 kgretainer
Japan white rabbit
Output current of CL biosensor
Calibration range
0.03~5.0mmol/l
glucose concentration (mmol/l)
outp
ut c
urre
nt (µ
A)
glucose concentration (mmol/l)
outp
ut c
urre
nt (µ
A)
Continuous monitoring of tear glucose
Ave. : 0.022 µA
estimated tear glu. : 0.13 mmol/l
(reported value 0.14 mmol/l)
Calibration curve for glucose Calibration curve for glucose
Calibration range
0.03~5.0mmol/l
before after
Before application After application
11
Glucose administration
Microscopic examination
Oral glucose tolerance test (OGTT)Comparison between tear and blood
No damage such as irritation
Tear fluid : Future potential as non-invasive sample for diabetes
continuous
before after
glucose administration
8min delay
dosage amount: 1g/weight-kg
Following blood change
blood
needleglucose
Tear glucose Blood glucose
fixator
Mouthguard biosensor for Saliva glucose mentoring
Saliva chemicals
glucoseascorbic acidureauric acidammoniacitric acid
mM
mMmMmMmM
μM
steady irritant
0.055 0.05610.4±4.3 4
3.22 2.170.09 0.18
0.01-0.060.004
(Data Book of Biochemistry 1980, Tokyo Kagaku Dojin) (Journal of the Japan Diabetes Society 1997;40)
Glucose b/w blood & saliva
Saliva monitoring for non‐invasive evaluation
12
Biosensor Mouthguard type biosensor
・enzyme immobilization・bonding to mouthguard・safety
Biocompatible polymer Mouthguard material
Use application・sports・teeth aligner・clenching protection・occlusal harmony
Shape forming sheet
Real time measurement of saliva glucose for non‐invasive blood glucose evaluation
glucose + O2 gluconolactone + H2O2GOD
Enzyme reactionSaliva glucose monitoringby mouthguard biosensor
GOD : glucose oxidase
Conclusions & Acknowledgements
Conclusions1) Soft CL biosensor was fabricated using Soft-MEMS
techniques and biocompatible polymers. # calibration range: 0.03 – 5.0 mmol/l# capable of continuous & real time tear sugar monitoring
2) Mouthguard telemetric biosensor was developed using hard dental material and BLE 4.0 tech.# detail and physiological results, soon.
AcknowledgementsResearch funding supports » JST A-STEP program» MEXT “Next-Generation Actuators
Leading Breakthroughs”» MEXT “Advanced Research Program in
Sensing Biology”.» JSPS Grant-in-Aid for Scientific Research» JSPS Grant-in-Aid for Exploratory» JST Tokyo-CREATE
Thank you !